Empty *_epi_field.nii file after using DIFFPREP

Submitted by jmollink on
Category

Hi Tortoise community,

 

After applying DIFFPREP to process ex-vivo tissue blocks, my final EPI corrected diffusion images should be stored in the EPI_field.nii. While the file is created, it contents remain empty.. I don't get any error messages in my terminal. Any ideas? Possibly the pipeline is a bit different when using non whole-brain data?

 

Cheers,

Jeroen

 

Submitted by EBHutch on Wed, 10/17/2018 - 10:57

Permalink

Jeroen -

I think it would solve your problem to use a registration settings file that is customized for your DWIs, especially for their spatial dimensions. 

Generally, the default settings for DIFFPREP assume your images are of an in-vivo human brain, but for pre-clinical or ex-vivo data we need to change the settings.  The flag to use in the DIFFPREP command is --reg_settings and the field should be the name for a file that exists in your */DIFF_PREP_WORK directory with the extension *.dmc

I'll put an example below for you to start with, you can copy it into a text file and save as <>.dmc within your DIFF_PREP_WORK directory. It is important that you change the fields of this file  (using any text editor) to suit your sample and registration.  I usually spend some time at the beginning of a study optimizing this file and then use it consistently for all samples.

 

Here is the registration settings content I used recently for a tissue block:

<!-- DTI initialization file -->
<!-- do not remove the above comment line -->

<!--***********************************************************************-->
<!-- CORRECTION MODE                                                       -->
<!--                                                                       -->
<!-- Specifies which undesirable effects will be corrected. Predefined     -->
<!-- motion & eddy distortion correction optimization settings (each       -->
<!-- setting points to a file in the software settings folder:             -->
<!--   off        :No DWI registration.                                    -->
<!--   rigid      :Corrects only motion.                                   -->
<!--   quadratic  :Eddy currents are modeled with quadratic functions.     -->
<!--               Quadratic                                               -->
<!--               model is sufficient 99% of the time.                    -->
<!--   cubic      :Eddy currents are modeled with cubic functions.         -->
<!--   multistart :Eddy currents are modeled with quadratic functions.     -->
<!--               The rigid search space is VERY LARGE. This can be used  -->
<!--               to register sagittal images to axial for example.       -->
<!-- You can also create a custom settings file in ~/DIFF_PREP_WORK and    -->
<!-- put its name here.                                                    -->
<correction_mode>rigid</correction_mode>


<!--***********************************************************************-->
<!-- EPI CORRECTION                                                        -->
<!--                                                                       -->
<!-- Predefined EPI distortion correction optimization settings. Each      -->
<!-- setting points to a file in the software settings folder.             -->
<!--                                                                       -->
<!--   off       :Do not perform EPI distortion correction.                -->
<!--   ITKBSP    :Perform BSplines transformation based correction with    -->
<!--              ITK. Similar to previous version of DIFFPREP.            -->
<!--   ITKBSPv4  :Perform BSplines transformation based correction with    -->
<!--              itkv4. Multiresolution. More powerful than ITKBSP but    -->
<!--              more sensitive.                                          -->
<!--   ANTSSyN   :Perform SyN transformation based correction with ANTs.   -->
<!--              Most powerful but most sensitive too.                    -->
<!-- The parameters for each method is described in their corresponding    -->
<!-- files.                                                                -->
<EPI_optimization_settings>off</EPI_optimization_settings>


<!--***********************************************************************-->
<!-- REFERENCE VOL INDEX                                                   -->
<!--                                                                       -->
<!-- Among possibly many b=0 s/mm2 images, the index of the b=0 image      --> 
<!-- (within only b=0 images) to be used as template. Starting index is    -->
<!-- zero, so that b0_id=2 will select the 3rd b=0 image. Defaults to the  -->
<!-- first b=0 image with id of 0.                                         -->
<b0_id>0</b0_id>
                
                
<!--***********************************************************************-->
<!-- GIBBS RINGING CORRECTION                                              -->
<!--                                                                       -->
<!-- For the original unprocessed DWIs. See Kellner, Dhital, Kiselev and   -->
<!-- Resiert, MRM 2016, 76:1574-1581.                                      -->
<!--   off       :Do not perform Gibbs ringing correction.                 -->
<!--   on        :Perform Gibbs ringing correction.                        -->
<gibbs_ringing_correction>off</gibbs_ringing_correction>


<!--***********************************************************************-->
<!-- DENOISING                                                             -->
<!--                                                                       -->
<!-- Methods for DWI images denoising based on J. Veraart, E. Fieremans,   -->
<!-- and D.S. Novikov, `Diffusion MRI noise mapping using random matrix    -->
<!-- theory' Magn Res. Med., early view (2016), doi: 10.1002/mrm.26059.    -->
<!--   off       :Still performs the denoising but only uses the output    -->
<!--              noise image for noise computations.                      -->
<!--   for_reg   :Uses the denoised DWIs for registration but outputs as   -->
<!--              the final images the corrected original images.          -->
<!--   for_final :Uses the denoised DWIs for both registration and as the  -->
<!--              final corrected images.                                  -->
<denoising>for_final</denoising> 


<!--***********************************************************************-->
<!-- HUMANITY                                                              -->
<!--                                                                       -->
<!-- Is it a human brain? Specialized processing is performed if the       -->
<!-- brain is human.                                                       -->
<!--   0         :No, not human brain.                                     -->
<!--   1         :Yes, human brain.                                        -->
<is_human_brain>0</is_human_brain>
                               
                               
<!--***********************************************************************-->
<!-- UPSAMPLING GEOMETRY                                                   -->
<!--                                                                       -->
<!-- Options.                                                              -->
<!--   off       :No upsampling.                                           -->
<!--   all       :Upsampling in all directions.                            -->
<!--   slice     :Upsampling in slice direction only.                      -->
<!-- If value entered does not match any of the above, it defaults to      -->
<!-- "all".                                                                -->
<upsampling>slice</upsampling>
                            
                            
<!--***********************************************************************-->
<!-- UPSAMPLING AMOUNT                                                     -->
<!--                                                                       -->
<!-- Upsampling factor.  By what factor each dimensions's resolution will  -->
<!-- be reduced (improved) by.                                             -->
<upsampling_factor>2.0000000</upsampling_factor>                           
                                                                          

<!--***********************************************************************-->
<!-- INTERPOLATION METHOD                                                  -->
<!--                                                                       -->
<!-- Determines the algorithm to use for upsampling.                       -->
<!--                                                                       -->
<!-- Will take effect only if <upsampling> is all or slice_only            -->
<!--   NN        :3D Nearest neighbor                                      -->
<!--   Linear    :Trilinear                                                -->
<!--   BSPCubic  :B-splines cubic.                                         -->
<!--   Sync      :Hamming sync                                             -->
<upsampling_method>Linear</upsampling_method> 
                                                
                                                
<!--***********************************************************************-->
<!-- HIGH B-VALUE REGISTRATION                                             -->
<!--                                                                       -->
<!-- High b-value registration for HARDI data. Only use this option if     --> 
<!-- your data have very high b-values which result in low SNR, as this    -->
<!-- option is slow.                                                       -->
<!-- Options for 'perform_high_bval_correction':                           -->
<!--   0         :Standard correction (i.e., no extra correction; default).-->
<!--   1         :Enable high b-value correction.                          -->
<!--                                                                       -->
<!-- If "1", then all volumes with bval <= 'high_b_dti_end_bval' will be   -->
<!-- corrected with standard processing. For vols with                     -->
<!-- 'high_b_dti_end_bval' < bval <= 'high_b_mapmri_start_bval',           -->
<!-- a noise- and distortion-free synthetic image will be synthesized from -->
<!-- the tensor model and will be used as the target.                      -->
<!-- For volumes with bval > 'high_b_mapmri_start_bval', a noise- and      -->
<!-- distortion-free synthetic image will be synthesized from the MAPMRI   -->
<!-- model and will be used as the target.                                 -->
<perform_high_bval_correction>0</perform_high_bval_correction>
<high_b_dti_end_bval>1800</high_b_dti_end_bval>
<high_b_mapmri_start_bval>3000</high_b_mapmri_start_bval>


<!--***********************************************************************-->
<!-- PRE-SMOOTHING                                                         -->
<!--                                                                       -->
<!-- Smoothing of images prior to registration. Specify whether or not to  -->
<!-- smooth the source and target images prior to registration (the final  -->
<!-- output, however, will not be smoothed). Options are:                  -->
<!--   off       :                                                         -->
<!--   gaussian  :                                                         -->
<!--   PM        :Perona-Malik.                                            -->
<smoothing>off</smoothing>
                          
              
                          
<!--***************************************************************************-->
<!--Place the center of mass of the image to the center pixel in the working FoV?-->
<center_of_mass>1</center_of_mass>
                     


<!--***************************************************************************-->
<!--Desired working FoV.  Applied to template image during cropping (value in mm)-->
<!--Only for internal processing. Does not affect the output.-->
<FoVx>32.0000</FOVx>
<FoVy>40.00000</FOVy>
<FoVz>6.50000</FOVz>


                               
                                  
<!--***********************************************************************-->
<!-- OUTPUT INTERMEDIATE, I                                                -->
<!--                                                                       -->
<!-- Write intermediate motion & eddy corrected data (the transformation   -->
<!-- file is written regardless). This should be written just for          -->
<!-- troubleshoothing purposes.                                            -->
<!--   0         :No, don't write intermediate files out.                  -->
<!--   1         :Yes, write intermediate files out.                       -->
<write_rpd>0</write_rpd>


<!--***********************************************************************-->
<!-- OUTPUT INTERMEDIATE, II                                               -->
<!--                                                                       -->
<!-- Write intermediate motion & eddy & EPI distortion corrected data (the -->
<!-- transformation file is written regardless). This should be written    -->
<!-- just for troubleshoothing purposes.                                   -->
<!--   0         :No, don't write intermediate files out.                  -->
<!--   1         :Yes, write intermediate files out.                       -->
<write_rpd_epi>0</write_rpd_epi>
                                
                                
<!--***********************************************************************-->
<!-- OUTPUT INTERMEDIATE, III                                              -->
<!--                                                                       -->
<!-- Keep intermediate files?:                                             --> 
<!--   0         :No, don't keep the intermediate files.                   -->
<!--   1         :Yes, keep the intermediate files.                        -->
<keep_intermediate_files>0</keep_intermediate_files>


<!--***********************************************************************-->
<!-- OUTPUT RESOLUTION                                                     -->
<!--                                                                       -->
<!-- Final diffusion weighted and structural image output resolution       -->
<!-- (value in mm).                                                        -->
<res_x>0.25</res_x>
<res_y>0.25</res_y>
<res_z>0.25</res_z>